Magnetic FT totaliser measure 50ltr of solution passed but actual solution was 42 ltr and it was matched with solution tank LT. So how I resolved this problem.Why FT measures max flow than actual…? Solution = Hydrogen peroxide
-
Remove it from the line, power up, and check the zero. It should be stable, if not then set zero as per the procedure given for the particular model/ brand
-
After verifying zero, make sure fluid parameters are properly entered and verify the K factor for your application. This calculation should be given in the manual.
-
For a magnetic flow meter you need laminar flow and the line must at-least half full
-
power supply and signal cables must be properly grounded to their own potentials. Mag flow meter is very sensitive for grounding.