# Why we set gap voltage -10 in vibration probe?

**URL:** <https://forumautomation.com/t/why-we-set-gap-voltage-10-in-vibration-probe/8857>\
**Category:** Field Instrumentation\
**Created:** [November 25, 2020, 5:03am UTC](https://forumautomation.com/t/why-we-set-gap-voltage-10-in-vibration-probe/8857 "2020-11-25T05:03:43Z")\
**Posts on this page:** 2\
**Page:** 1

<div class="post-metadata">

**Author:** ![instrumentationtechi](https://forumautomation.com/letter_avatar_proxy/v4/letter/i/50afbb/32.png) [@instrumentationtechi](https://forumautomation.com/u/instrumentationtechi)\
**Post date:** [November 25, 2020, 5:03am UTC](https://forumautomation.com/t/why-we-set-gap-voltage-10-in-vibration-probe/8857/1 "2020-11-25T05:03:43Z")

</div>

Explain why the voltage is set to -10 in a vibration probe

---

<div class="post-metadata">

**Author:** ![ashlin](https://forumautomation.com/user_avatar/forumautomation.com/ashlin/32/3775_2.png) [@ashlin](https://forumautomation.com/u/ashlin)\
**Post date:** [November 25, 2020, 5:23am UTC](https://forumautomation.com/t/why-we-set-gap-voltage-10-in-vibration-probe/8857/2 "2020-11-25T05:23:02Z")

</div>

The Gap is usually set at -10V. The reason for setting at -10 volt is, proximitor allows the probe to work from -2 volt to -18 volt. So the voltage difference of 2 to 10 and 10 to 18 is 8 volt. When setting at -10 volt, the probe can equally measure vibration at both sides, That is when the shaft is going away from a probe or towards the probe.