# Procedure for performing ICE bath test bath of standard temperature reference probe

**URL:** <https://forumautomation.com/t/procedure-for-performing-ice-bath-test-bath-of-standard-temperature-reference-probe/237>\
**Category:** Temperature Measurement\
**Created:** [February 6, 2015, 5:37pm UTC](https://forumautomation.com/t/procedure-for-performing-ice-bath-test-bath-of-standard-temperature-reference-probe/237 "2015-02-06T17:37:36Z")\
**Posts on this page:** 1\
**Page:** 1

<div class="post-metadata">

**Author:** ![denizen](https://forumautomation.com/letter_avatar_proxy/v4/letter/d/848f3c/32.png) [@denizen](https://forumautomation.com/u/denizen)\
**Post date:** [February 6, 2015, 5:37pm UTC](https://forumautomation.com/t/procedure-for-performing-ice-bath-test-bath-of-standard-temperature-reference-probe/237/1 "2015-02-06T17:37:37Z")

</div>

ICE bath test is used for testing standard temperature reference probe .This test will show if the standard probe has drifted. Used in conjunction with a dry block calibrator for Temperature calibrations over the range of -30°C to +140°C.

Things required for performing ICE bath calibration - Flask,ICE,Standard Reference Probe 1.Crush the ICE and fill the flask. 2.The flask will be allowed to stabilize for a few minutes until the ice looks uniformly grey 3.Immerse the probe in the flask and stabilizing until the reading is stable to within 0.1°C. (Do not allow probe to touch the bottom of the flask) 4.Record reading.

Accepted error is +/- 0.2 deg C