# How to use Dry Block Calibrator for Temperature calibration

**URL:** <https://forumautomation.com/t/how-to-use-dry-block-calibrator-for-temperature-calibration/236>\
**Category:** Calibration\
**Created:** [February 6, 2015, 5:20pm UTC](https://forumautomation.com/t/how-to-use-dry-block-calibrator-for-temperature-calibration/236 "2015-02-06T17:20:28Z")\
**Posts on this page:** 1\
**Page:** 1

<div class="post-metadata">

**Author:** ![denizen](https://forumautomation.com/letter_avatar_proxy/v4/letter/d/848f3c/32.png) [@denizen](https://forumautomation.com/u/denizen)\
**Post date:** [February 6, 2015, 5:20pm UTC](https://forumautomation.com/t/how-to-use-dry-block-calibrator-for-temperature-calibration/236/1 "2015-02-06T17:20:29Z")

</div>

Dry Block Calibrators are used for temperature calibration.

Equipment required : Dry Block calibrator,Reference probe UUT - Unit under test , Temperature sensor to be calibrated

1. Check the tolerance of reference probe using ice bath 2.Insert the reference probe and UUT together into suitable bores 3.Set the required temperature 4.Wait 6 minute or until the indication of the standard probe is stable to within ±0.1°C. 5 Record the following temperature Reference probe,UUT,Ambient 6 Repeat the step for different temperature